[mcstas-users] Reflectivity and transmission files of Ge crystal
lucia.alianelli at diamond.ac.uk
lucia.alianelli at diamond.ac.uk
Fri May 11 16:52:17 CEST 2018
Dear All,
Please ignore if you are not interested in imperfect crystal (monochromators, analysers etc) computations.
I am sending a report written on the calculation of mosaic / bent / gradient crystal calculations a long time ago. This was part of a PhD thesis project - like the paper you discussed in this thread: J Appl Crys 37 732+ (http://dx.doi.org/10.1107/S0021889804015390)
The reflectivity code is included at the end of the report - unfortunately I could not find the Fortran version in my old files - but I’m sure it should be possible to translate the code easily into any other language if of interest.
Feel free to send feedback
Regards
Lucia
Lucia Alianelli
Senior Optics Scientist
Diamond Light Source Ltd
From: mcstas-users [mailto:mcstas-users-bounces at mcstas.org] On Behalf Of ????
Sent: 11 May 2018 10:36
To: IversonErikB.; pkwi@; saed@
Cc: mcstas-users at mcstas.org
Subject: Re: [mcstas-users] Reflectivity and transmission files of Ge crystal
Dear Prof. Erik
Thank you for your reply and for these valuable comments, I really appreciate it
Let me start from point #3 and 4 first;
HANARO-HRPD is installed on thermal beam tube with Maxwellian spectrum, where the capture flux and particle flux are equal by definition as I know. And the input Temp in my code is 37 degree not kelvin and the code will change it to 310 K while running,
Point #2; I have checked the link that you sent, but the research paper investigates the peak reflectivity for (Ge 311) which has high integrated reflectivity comparing to (Ge 331) that used in HANARO-HRPD, so I think I shall not use a peak reflectivity of 0.5 ~ 0.6 for 1.836 AA for (Ge 331), are you agree with me or there is something I miss?
Point #1: I tried to find some data about the wavelength spread of Ge monochromator at HANARO-HRPD but unfortunately I couldn't find!
But I'm wondering about this point , therefore let's assume that the measured wavelength spread is much larger than the calculated one by McStas, so what's next ?
The wavelength spread is governed mainly by mosaicity of Monochromator, so in order to match the calculated wavelength spread (McStas) with the measured one I have to change the mosaic angle in McStas code which is a characteristic property of Mononchromator that should not be changed at all, am I wrong?!
I wonder if there is another way to match them without changing the parameters of Monochromator.
Thank you again for your efforts and looking forward to your response
Best Regards
Saed Almomani
[Image removed by sender.]
-----Original Message-----
보낸사람(From) : Iverson Erik B. <iversoneb at ornl.gov>
받는사람(To) : pkwi at fysik.dtu.dk <pkwi at fysik.dtu.dk>;saed at kaeri.re.kr <saed at kaeri.re.kr>
참조(Cc) : Iverson Erik B. <iversoneb at ornl.gov>;mcstas-users at mcstas.org <mcstas-users at mcstas.org>
보낸일자(Sent) : 2018-05-10 22:32:03
메일제목(Subject) : Re: [mcstas-users] Reflectivity and transmission files of Ge crystal
Dear Saed,
At first glance, I think there are four things for you to check. You say that your value at the monochromator is what you expect, but the sample is down by a factor of eight - as you indicated earlier, this definitely suggests issues with the monochromator (germanium) modeling.
1) Your measured flux on sample is 1.3E6 n/cm2/s at 1.836 AA. This is a little bit incomplete for diagnostic purposes- what is the wavelength spread? What is the divergence? If you can compare those bounds to that produced by the calculation, you may find something informative. For example, it may be that your measured wavelength spread is larger than your modeled spread, and thus the calculation is missing something.
2) You use a reflectivity of 0.3 in the model (for 1.836 AA off of 331). While that will depend significantly on the thickness of the crystal as well as other factors, that seems low to me. Look at J Appl Crys 37 732+ (http://dx.doi.org/10.1107/S0021889804015390). There, they report peak reflectivity measurements of order 0.5 to 0.6, with a reflectivity FWHM half again larger than the intrinsic mosaic.
While not related to the monochromator per se, it is also worth asking:
3) Just to be paranoid - you quote a flux integral of 2.7E8 at the monochromator position - is that particle flux or capture flux? You're on a cold source (at least the Maxwellian temperature of 37 suggests that), so the capture flux is numerically significantly higher than the particle flux integral. You've almost certainly got the right value - but it's easy enough to check that I claim it's worth checking. (The same would technically apply to the beam tube flux of 1.28E14 - if that's a measured capture flux, it's a specific factor higher than the particle flux, and that same factor would apply at the monochromator position.)
4) Most complicated - hydrogen cold sources don't necessarily have a spectrum well fit by a Maxwellian, and your sample flux really only runs just around that one wavelength. It's easy for the integral to match with the value at one wavelength to be significantly different. Are there spectral measurements that support a 37-K Maxwellian?
Good luck,
Erik
On Thu, 2018-05-10 at 16:43 +0900, 새드하셈 wrote:
Dear Dr. Peter
First of all thank you so much for your quick response, I appreciate your help and effort
Regarding your request, you can find the attached file of my McStas code for the HANARO-HRPD with this email.
I would like to mention that I wrote almost all required descriptions and information as comments inside the code itself.
I'm sure that you can follow the design without any conflict. But in case you need more information I will response quickly.
Thank you in advance for your efforts and cooperation
[Image removed by sender.]
-----Original Message-----
보낸사람(From) : Peter Kjær Willendrup <pkwi at fysik.dtu.dk>
받는사람(To) : 새드하셈 <saed at kaeri.re.kr>
참조(Cc) : mcstas-users at mcstas.org <mcstas-users at mcstas.org>
보낸일자(Sent) : 2018-05-09 22:56:02
메일제목(Subject) : Re: [mcstas-users] Reflectivity and transmission files of Ge crystal
Dear Saed,
On 9 May 2018, at 11:30 , 새드하셈 <saed at kaeri.re.kr<mailto:saed at kaeri.re.kr>> wrote:
Actually I have one question and one inquiry:
Q: I'm trying to simulate the HANARO_HRPD by using McStas software. But after I inserted all required components and their parameters as designed/as they are on ground now, I found a kind of difference between the McStas_flux magnitude and the Real_flux magnitude both at sample position in (n/cm2.s) !! and the difference is high enough to say there is something wrong I can't figure it ! {Real_flux is around (1.3e6) n/cm2.s where McStas_flux is (1.9e5) n/cm2.s.} although I get almost exact matching result at monochromator position.
It is always easier to comment or correct if there is acces to the instrument file, as well as other necessary data.
I would therefor encourage you to send the instrument file, plus descriptive material about your beamline in response to this email.
And my concerns also increased when some experts (2 experts to be honest) told me that McStas doesn't show the flux magnitude as calculated or expected !! So is this True ?! and if yes, what is the accepted difference between McStas and Real, i.e. the criteria or error margin to consider?
I my experience McStas is reasonably reliable, and to a large extent simulates exactly what one parametrises. This of course means that if the source emission or certain material- or geometrical data are put in with values that are off, the simulation will naturally also be off. :-)
A rule of thumb is that with modest effort, one can arrive within 10% of what one would measure. Better agreement than that typically requires lots of more thinking, coding, simulating and benchmarking.
But again, without access to what you are trying to simulate and how, it is hard to judge what is going on. We are "flying blind".
Inquiry: I tried to get the Reflectivity files of Ge crystal to define a curved Monochromator with four wavelengths (1.54, 1.834, 1.866, 2.224 A), but unfortunately I couldn't find any source that provide a tabulated data of the wavelength and the absolute Ge crystal reflectivity. So is there anyone who could help me in this also?
I must admit I am not much of a monochromator-expert, but I would try looking among papers by Freund et. al from the 1970’s and 80’s.
Best regards,
Peter Willendrup
Peter Kjær Willendrup
Forskningsingeniør, Speciakonsulent
Næstformand for DTU Fysik LSU
DTU Physics
[Image removed by sender.]
Technical University of Denmark
[Image removed by sender.]
Department of Physics
Fysikvej
Building 307
DK-2800 Kongens Lyngby
Direct +45 2125 4612
Mobil +45 2125 4612
Fax +45 4593 2399
pkwi at fysik.dtu.dk<mailto:pkwi at fysik.dtu.dk>
_______________________________________________
mcstas-users mailing list
mcstas-users at mcstas.org<mailto:mcstas-users at mcstas.org>
https://mailman2.mcstas.org/mailman/listinfo/mcstas-users
--
Erik B. Iverson, PhD. Spallation Neutron Source
Building 8600, MS6466 1 (865) 241-6970
Oak Ridge National Laboratory 1 (865) 574-4140 (FAX)
Oak Ridge, TN 37831 iversoneb at ornl.gov<mailto:iversoneb at ornl.gov>
--
This e-mail and any attachments may contain confidential, copyright and or privileged material, and are for the use of the intended addressee only. If you are not the intended addressee or an authorised recipient of the addressee please notify us of receipt by returning the e-mail and do not use, copy, retain, distribute or disclose the information in or attached to the e-mail.
Any opinions expressed within this e-mail are those of the individual and not necessarily of Diamond Light Source Ltd.
Diamond Light Source Ltd. cannot guarantee that this e-mail or any attachments are free from viruses and we cannot accept liability for any damage which you may sustain as a result of software viruses which may be transmitted in or with the message.
Diamond Light Source Limited (company no. 4375679). Registered in England and Wales with its registered office at Diamond House, Harwell Science and Innovation Campus, Didcot, Oxfordshire, OX11 0DE, United Kingdom
-------------- next part --------------
An HTML attachment was scrubbed...
URL: <http://mailman2.mcstas.org/pipermail/mcstas-users/attachments/20180511/6b36a07a/attachment-0001.html>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ~WRD000.jpg
Type: image/jpeg
Size: 823 bytes
Desc: ~WRD000.jpg
URL: <http://mailman2.mcstas.org/pipermail/mcstas-users/attachments/20180511/6b36a07a/attachment-0002.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: image001.jpg
Type: image/jpeg
Size: 332 bytes
Desc: image001.jpg
URL: <http://mailman2.mcstas.org/pipermail/mcstas-users/attachments/20180511/6b36a07a/attachment-0003.jpg>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: ILL03AL05T.pdf
Type: application/pdf
Size: 273327 bytes
Desc: ILL03AL05T.pdf
URL: <http://mailman2.mcstas.org/pipermail/mcstas-users/attachments/20180511/6b36a07a/attachment-0001.pdf>
More information about the mcstas-users
mailing list